Share Your Feedback – Help Us Improve Search on Community!
Please take a few minutes to participate in our Search Feedback Survey. Your insights will help us deliver the results you need faster and more accurately.
Click here to take the survey
Procedure to switch between Single Device and Network modes in ION Setup
Metering & Power Quality
Schneider Electric support forum about Power Meters (ION, PowerTag, PowerLogic) and Power Quality from design, implementation to troubleshooting and more.
Search in
Improve your search experience:
Exact phrase→Use quotes " "(e.g., "error 404")
Wildcard→Use * for partial words(e.g., build*, *tion)
AND / OR→Combine keywords(e.g., login AND error, login OR sign‑in)
Keep it short→Use 2–3 relevant words, not full sentences
Filters→Narrow results by section(Knowledge Base, Users, Products)
Send a co-worker an invite to the portal.Just enter their email address and we'll connect them to register. After joining, they will belong to the same company.
You have entered an invalid email address. Please re-enter the email address.
This co-worker has already been invited to the Exchange portal. Please invite another co-worker.
Please enter email address
Send InviteCancel
Invitation Sent
Your invitation was sent.Thanks for sharing Exchange with your co-worker.
Link copied. Please paste this link to share this article on your social media post.
2013-02-2811:09 AM
Procedure to switch between Single Device and Network modes in ION Setup
For temporary purposes, you can perform the following procedure:
Select from the menu, File -> Exit and when the Exit Options dialog is displayed, select the "User Logoff" radio button and click on OK. This will display the standard logon dialog.
Use the default logon of "supervisor" and the default password of 0 (zero). To access single device mode, make sure the "Single ION device configuration mode" checkbox is checked, otherwise network mode will be used. Click on OK to proceed.
To change the ION Setup start-up mode permamently:
Select from the menu, Tools -> Options (or click on the checkbox icon on the toolbar) and click on the View tab.
Select a new Startup Mode as either "Metershop Mode" (Single Device), "Network Mode", or "Power User Mode" (shows the standard logon dialog at startup which will allow you to switch between the modes but requires you to enter a password each time).
Click on OK to save the changes.
Reminder that Single Device mode will only function with direct serial/modem or TCP connections (no gateway connections) and will only function with ION devices which support ION protocol (no 6200 ION or CM/PM PowerLogic devices which rely on Modbus).