PM8000 Frequency resolution

Metering & Power Quality

Schneider Electric support forum about Power Meters (ION, PowerTag, PowerLogic) and Power Quality from design, implementation to troubleshooting and more.

cancel
Turn on suggestions
Auto-suggest helps you quickly narrow down your search results by suggesting possible matches as you type.
Showing results for 
Show  only  | Search instead for 
Did you mean: 
  • Home
  • Schneider Electric Community
  • EcoStruxure Power & Grid
  • Metering & Power Quality
  • PM8000 Frequency resolution
Options
  • Mark Topic as New
  • Mark Topic as Read
  • Float this Topic for Current User
  • Bookmark
  • Subscribe
  • Mute
  • Printer Friendly Page
Invite a Co-worker
Send a co-worker an invite to the portal.Just enter their email address and we'll connect them to register. After joining, they will belong to the same company.
You have entered an invalid email address. Please re-enter the email address.
This co-worker has already been invited to the Exchange portal. Please invite another co-worker.
Please enter email address
Send Invite Cancel
Invitation Sent
Your invitation was sent.Thanks for sharing Exchange with your co-worker.
Send New Invite Close
Top Experts
User Count
Charles_Murison
Picard Charles_Murison Picard
477
Ramasamy_N
Captain Ramasamy_N Captain
113
Robert_Lee
Admiral Robert_Lee Admiral
92
Mehran_Mehrnia
Captain Mehran_Mehrnia Captain
85
View All

Invite a Colleague

Found this content useful? Share it with a Colleague!

Invite a Colleague Invite
Solved Go to Solution
Back to Metering & Power Quality
Solved
ausfrosty
Crewman ausfrosty
Crewman

Posted: ‎2022-08-28 06:10 PM

0 Likes
3
2148
  • Mark as New
  • Bookmark
  • Subscribe
  • Mute
  • Permalink
  • Print
  • Email to a Friend
  • Report Inappropriate Content

Link copied. Please paste this link to share this article on your social media post.

‎2022-08-28 06:10 PM

PM8000 Frequency resolution

Hi All,

 

I am trying to determine if the PM8000 meets the requirements of the AEMO Mareket Ancillary Service Specificaiton Verion 7.1 in terms of Frequency resolution for Fast FCAS? Table 4 in 5.3.2  

https://aemo.com.au/-/media/files/stakeholder_consultation/consultations/nem-consultations/2021/mass...

 

As per the attached screen shot the frequency resolution required is <= 0.0025Hz (2.5mHz) for Fast FCAS.

 

If it does not is there a Schneider meter that does?

 

Attachments
Labels
  • Labels:
  • PM8000
Tags (1)
  • Tags:
  • english
Reply
  • All forum topics
  • Previous Topic
  • Next Topic

Accepted Solutions
DanL
Captain DanL Captain
Captain

Posted: ‎2022-08-29 09:47 AM

1 Like
2
2127
  • Mark as New
  • Bookmark
  • Subscribe
  • Mute
  • Permalink
  • Print
  • Email to a Friend
  • Report Inappropriate Content

Link copied. Please paste this link to share this article on your social media post.

‎2022-08-29 09:47 AM

Yes, the frequency registers have better than 2.5mHz precision on the PM8000 (when read or recorded in Float32 format).

The PM8000 uses a 40MHz clock to measure the periods of the zero crossings of the fundamental voltage.

 

Model ION Module ION Register Measurement window Update period Standard Compliance Claimed Accuracy Precision
PM8000 Power Quality Aggregator Power Frequency 10 s 10 s

IEC61000-4-30

Class S*

+/-10mHz

< 0.01mHz

(7 digits)

PM8000 Power Meter Module Freq 1 s 1 s

IEC61557-12

CL,02

+/-0.02% = +/-10mHz

< 0.01mHz

(7 digits)

PM8000 HS Power Meter Module HS Freq 1 cycle 1/2 cycle    

< 0.1mHz

(6 digits)

 

See Answer In Context

Tags (1)
  • Tags:
  • english
Reply
Replies 3
DanL
Captain DanL Captain
Captain

Posted: ‎2022-08-29 09:47 AM

1 Like
2
2128
  • Mark as New
  • Bookmark
  • Subscribe
  • Mute
  • Permalink
  • Print
  • Email to a Friend
  • Report Inappropriate Content

Link copied. Please paste this link to share this article on your social media post.

‎2022-08-29 09:47 AM

Yes, the frequency registers have better than 2.5mHz precision on the PM8000 (when read or recorded in Float32 format).

The PM8000 uses a 40MHz clock to measure the periods of the zero crossings of the fundamental voltage.

 

Model ION Module ION Register Measurement window Update period Standard Compliance Claimed Accuracy Precision
PM8000 Power Quality Aggregator Power Frequency 10 s 10 s

IEC61000-4-30

Class S*

+/-10mHz

< 0.01mHz

(7 digits)

PM8000 Power Meter Module Freq 1 s 1 s

IEC61557-12

CL,02

+/-0.02% = +/-10mHz

< 0.01mHz

(7 digits)

PM8000 HS Power Meter Module HS Freq 1 cycle 1/2 cycle    

< 0.1mHz

(6 digits)

 

Tags (1)
  • Tags:
  • english
Reply
ausfrosty
Crewman ausfrosty
Crewman

Posted: ‎2022-08-29 03:58 PM

0 Likes
1
2119
  • Mark as New
  • Bookmark
  • Subscribe
  • Mute
  • Permalink
  • Print
  • Email to a Friend
  • Report Inappropriate Content

Link copied. Please paste this link to share this article on your social media post.

‎2022-08-29 03:58 PM

Hi Dan,

 

Thanks for that. Just what I needed.

 

I was wondering where you sourced the information from? Is there any official Schneider documentation with that written down?

I will likely need to show the client that the meter satifies the requirement. If it is not in official documentation an email from local Schneider team will hopefully suffice.

 

Thanks again.

Tags (1)
  • Tags:
  • english
Reply
Charles_Murison
Picard Charles_Murison Picard
Picard

Posted: ‎2022-08-30 09:37 AM

0 Likes
0
2095
  • Mark as New
  • Bookmark
  • Subscribe
  • Mute
  • Permalink
  • Print
  • Email to a Friend
  • Report Inappropriate Content

Link copied. Please paste this link to share this article on your social media post.

‎2022-08-30 09:37 AM

Hello @ausfrosty 

 

Standard Compliance and Claimed Accuracy can be found in the technical datasheet
https://www.se.com/ca/en/download/document/PLSED310058EN_Web/

 

Regards,

Charles

L4 Prime for Advanced metering and Utilities
Tags (1)
  • Tags:
  • english
Reply
Preview Exit Preview

never-displayed

You must be signed in to add attachments

never-displayed

 
To The Top!

Forums

  • APC UPS Data Center Backup Solutions
  • EcoStruxure IT
  • EcoStruxure Geo SCADA Expert
  • Metering & Power Quality
  • Schneider Electric Wiser

Knowledge Center

Events & webinars

Ideas

Blogs

Get Started

  • Ask the Community
  • Community Guidelines
  • Community User Guide
  • How-To & Best Practice
  • Experts Leaderboard
  • Contact Support
Brand-Logo
Subscribing is a smart move!
You can subscribe to this board after you log in or create your free account.
Forum-Icon

Create your free account or log in to subscribe to the board - and gain access to more than 10,000+ support articles along with insights from experts and peers.

Register today for FREE

Register Now

Already have an account? Login

Terms & Conditions Privacy Notice Change your Cookie Settings © 2025 Schneider Electric

This is a heading

With achievable small steps, users progress and continually feel satisfaction in task accomplishment.

Usetiful Onboarding Checklist remembers the progress of every user, allowing them to take bite-sized journeys and continue where they left.

of