Continue
Help
  • Explore Community
  • Get Started
  • Ask the Community
  • How-To & Best Practices
  • Contact Support
Notifications
Login / Register
Schneider Electric
Community
Community
Notifications
close
  • Categories
  • My Subscriptions
  • Forums
  • Knowledge Center
  • Blogs
  • Ideas
  • Events
Help
Help
  • Explore Community
  • Get Started
  • Ask the Community
  • How-To & Best Practices
  • Contact Support
Login / Register

Contact Support

Close

Ask our Experts

Have a question related to our products, solutions or services? Get quick support on community Forums

Email Us

For Community platform-related support, please email us

New Community Ranking System
Our Community ranking system has recently been updated. You may notice changes in user rankings and receive system messages or notifications. If you have questions about how the new ranking works, please refer to the announcement post for more details (click here).

ION8650 151 Errors

Metering & Power Quality

Schneider Electric support forum about Power Meters (ION, PowerTag, PowerLogic) and Power Quality from design, implementation to troubleshooting and more.

Search in

Improve your search experience:

  • Exact phrase → Use quotes " " (e.g., "error 404")
  • Wildcard → Use * for partial words (e.g., build*, *tion)
  • AND / OR → Combine keywords (e.g., login AND error, login OR sign‑in)
  • Keep it short → Use 2–3 relevant words , not full sentences
  • Filters → Narrow results by section (Knowledge Base, Users, Products)
cancel
Turn on suggestions
Auto-suggest helps you quickly narrow down your search results by suggesting possible matches as you type.
Showing results for 
Show  only  | Search instead for 
Did you mean: 
  • Home
  • Schneider Electric Community
  • EcoStruxure Power & Grid
  • Metering & Power Quality
  • ION8650 151 Errors
Options
  • Mark Topic as New
  • Mark Topic as Read
  • Float this Topic for Current User
  • Bookmark
  • Subscribe
  • Mute
  • Printer Friendly Page
Invite a Co-worker
Send a co-worker an invite to the portal.Just enter their email address and we'll connect them to register. After joining, they will belong to the same company.
You have entered an invalid email address. Please re-enter the email address.
This co-worker has already been invited to the Exchange portal. Please invite another co-worker.
Please enter email address
Send Invite Cancel
Invitation Sent
Your invitation was sent.Thanks for sharing Exchange with your co-worker.
Send New Invite Close
Top Experts
User Count
Charles_Murison
Charles_Murison Champion
581
Mehran_Mehrnia
Mehran_Mehrnia Champion
216
Ramasamy_N
Ramasamy_N Champion
117
View All

Recommended Forums

  • SmartConnector Forum

Invite a Colleague

Found this content useful? Share it with a Colleague!

Invite a Colleague Invite
Solved Go to Solution
Back to Metering & Power Quality
Start a Topic
Solved
Michael_Neas
Michael_Neas Troubleshooter
Troubleshooter

Posted: ‎2014-06-05 08:56 AM

0 Likes
1
1527
  • Mark as New
  • Bookmark
  • Subscribe
  • Mute
  • Permalink
  • Print
  • Email to a Friend
  • Report Inappropriate Content

Link copied. Please paste this link to share this article on your social media post.

‎2014-06-05 08:56 AM

ION8650 151 Errors

Hello,

Does anyone have additional information on 151 errors.  Specifically, what causes these errors on a meter, and what does a 151 error mean? 

I have a customer who was getting these on a couple meters.  When they started coming into the event log, they said that their load profile log stopped updating.  Sometimes a power cycle would correct the issue for a little while but the errors would come back with the same problem.

Thanks for the information in advance.

Michael Neas

Reply
  • All forum topics
  • Previous Topic
  • Next Topic

Accepted Solutions
sesa56545
sesa56545 Leader
Leader

Posted: ‎2014-06-05 09:53 AM

0 Likes
0
1475
  • Mark as New
  • Bookmark
  • Subscribe
  • Mute
  • Permalink
  • Print
  • Email to a Friend
  • Report Inappropriate Content

Link copied. Please paste this link to share this article on your social media post.

‎2014-06-05 09:53 AM

Michael

The short answer is one or more blocks in the flash can not be read.

UPGFileCantRead

As you said it recovers for a time after a power cycle so this condition would likley be resolved by a firmware upgrade to V405 or V406

If a firmware upgrade does not resolve the issue a Flash block is stuck in the read state and the meters would need to be returned for repair.

Root Cause summary:

The firmware design team working with the flash manufacturer discovered there was a potential that certain aspects of the meter's code might affect the internal memory management of the meter.  Firmware enhancements were incorporated into V405 to prevent the possibility of the flash read write issues from occurring.

Recommendations for the customer:

Should a similar occurrence happen the device can sometimes be corrected by power cycling.

If possible an upgrade to 8650V405 firmware which was enhanced to prevent these errors from occurring is suggested.

For devices that already exhibit the symptoms and do not return to normal operation after a power cycle the new firmware may not directly resolve the issue.

If the device has the new firmware and has been power cycled and continues to show the symptoms it will need to be returned for repair.

Schneider Electric internal corrective actions :

Improvements to the firmware have been identified through collaboration between Schneider Electric and its suppliers.

This new firmware will resume read/write cycles once power has been restored to the meter, will change memory blocks from an unknown state to a known state, and will improve memory reliability to safeguard against data loss.

See Answer In Context

Reply
Reply 1
sesa56545
sesa56545 Leader
Leader

Posted: ‎2014-06-05 09:53 AM

0 Likes
0
1475
  • Mark as New
  • Bookmark
  • Subscribe
  • Mute
  • Permalink
  • Print
  • Email to a Friend
  • Report Inappropriate Content

Link copied. Please paste this link to share this article on your social media post.

‎2014-06-05 09:53 AM

Michael

The short answer is one or more blocks in the flash can not be read.

UPGFileCantRead

As you said it recovers for a time after a power cycle so this condition would likley be resolved by a firmware upgrade to V405 or V406

If a firmware upgrade does not resolve the issue a Flash block is stuck in the read state and the meters would need to be returned for repair.

Root Cause summary:

The firmware design team working with the flash manufacturer discovered there was a potential that certain aspects of the meter's code might affect the internal memory management of the meter.  Firmware enhancements were incorporated into V405 to prevent the possibility of the flash read write issues from occurring.

Recommendations for the customer:

Should a similar occurrence happen the device can sometimes be corrected by power cycling.

If possible an upgrade to 8650V405 firmware which was enhanced to prevent these errors from occurring is suggested.

For devices that already exhibit the symptoms and do not return to normal operation after a power cycle the new firmware may not directly resolve the issue.

If the device has the new firmware and has been power cycled and continues to show the symptoms it will need to be returned for repair.

Schneider Electric internal corrective actions :

Improvements to the firmware have been identified through collaboration between Schneider Electric and its suppliers.

This new firmware will resume read/write cycles once power has been restored to the meter, will change memory blocks from an unknown state to a known state, and will improve memory reliability to safeguard against data loss.

Reply

You’ve reached the end of your document

WHAT’S NEXT?

Ask our Experts

Didn't find what you are looking for? Ask our experts!

My Dashboard

Check out the new Feeds and activities that are relevant to you.

To The Top!

Forums

  • APC UPS Data Center Backup Solutions
  • EcoStruxure IT
  • EcoStruxure Geo SCADA Expert
  • Metering & Power Quality
  • Schneider Electric Wiser

Knowledge Center

Events & webinars

Ideas

Blogs

Get Started

  • Ask the Community
  • Community Guidelines
  • Community User Guide
  • How-To & Best Practice
  • Experts Leaderboard
  • Contact Support

    Ask our Experts

    Have a question related to our products, solutions or services? Get quick support on community Forums

    Email Us

    For Community platform-related support, please email us

Subscribing is a smart move!
You can subscribe to this board after you log in or create your free account.
Forum-Icon

Create your free account or log in to subscribe to the board - and gain access to more than 10,000+ support articles along with insights from experts and peers.

Register today for FREE

Register Now

Already have an account? Login

Terms & Conditions Privacy Notice Change your Cookie Settings © 2026 Schneider Electric

Welcome!

Welcome to your new personalized space.

of

Explore