Did some one test the configuration of energy pulse ouptput of PM5111 with front panel device HIM?
It seems it is not possible to set the pulse output with the PM5111 HMI. When I set the device and save the modifications, the modifications are not saved
To do this on my product I have to use ION SETUP software.
Thanks for any feed back.
Philippe
Posted: 2015-05-25 01:08 AM
Link copied. Please paste this link to share this article on your social media post.
Hi Philipe Michon,
For present firmware (1.4.7 for MID meters, and 1.3.7 for non-MID meters), it can only be configured through ION Setup.
The factory has released a new firmware 1.6.0, which provides the possibility to configure through HMI. This new firmware version file will be available at site and Shopping Kiosk by this week.
Link copied. Please paste this link to share this article on your social media post.
Hello Chen,
thanks a lot for your feed back. Just on more precision:
Is there only PM51xx affected by this issue or PM53xx and PM55xx have the same behaviour?
Best regards.
Philippe
Posted: 2015-05-26 02:08 AM
Link copied. Please paste this link to share this article on your social media post.
Hello Philipe Michon,
It is applicable for PM5111, PM5331 and PM5341.
For PM5561, no this type of issue.