Welcome to the new Schneider Electric Community

It's your place to connect with experts and peers, get continuous support, and share knowledge.

  • Explore the new navigation for even easier access to your community.
  • Bookmark and use our new, easy-to-remember address (community.se.com).
  • Get ready for more content and an improved experience.

Contact SchneiderCommunity.Support@se.com if you have any questions.

Close
Invite a Co-worker
Send a co-worker an invite to the Exchange portal.Just enter their email address and we’ll connect them to register. After joining, they will belong to the same company.
Send Invite Cancel
84793members
354241posts

IMD-IT-S100-H Compliance with IEC 61558-2-15

Solutions for your Business Forum

No matter if your business is in the retail sector, food manufacturing, or healthcare, you'll find resources and a community of experts to discuss with on content tailored to your industry sector. Subscribe today.

SESA214337
Crewman Crewman
Crewman
0 Likes
0
594

IMD-IT-S100-H Compliance with IEC 61558-2-15

Dear My Colleagues, 

 

I am checking the IMD-IT-S100-H (IT Medical Transformers - Asia Range) specifications for medical application.

 

1. IMD - IT - S100H complied with IEC 61558-2-15: 2011 standard.

 

2. Clause 18.102 IEC 61558-2-15: 2011 standard mentions: The leakage current through the protective earth conductor from the enclosure to earth, shall not exceed 0,5 mA when measured at room temperature under no-load condition with the transformer supplied at the rated voltage and the rated frequency.

 

3. But on the transformer spec, leakage current item it shows that " The leakage current is less than 3.5mA between primary and secondary, primary and bracket, when input 230V/ 50HZ"

 

With the specification in [3], customer understands that "The leakage current < 3.5mA" and it didn't comply with item [2].

It is so confuse for me and also customer.

 

My question: Can we confirm with customer that: “our isolation transformer for medical locations IMD-IT-S100-H has leakage current through the protective earth conductor from the enclosure to earth is not exceed 0.5mA as per IEC 61558-2-15 standard ?"

 

If yes, do we have any update document to show it.

 

Thank you very much for your support.

 

 

 

 

 